3.8.5 Example Calculation Involving Intrinsic Viscosity
Let's say we are interested in a fast, approximate molecular weight of a polystyrene sample. We dissolve 0.10 g of the polymer in 100 ml of butanone and measure the flow times at 25°C in an Ubbelhode capillary viscometer. The results are as follows:
Pure butanone: 110 s
Polystyrene solution: 140 s
Starting with equation (3.84) and noting that the flow time is proportional to the viscosity, we have:
T_rel = 140/110 = 1.273
Using equation (3.86) and assuming that the concentration is near zero, we can approximate [n] as 2.73 x 10^-2 ml/g. This eliminates the need for extrapolation in Figure 3.14 for more accurate results.
Using the Mark-Houwink-Sakurada relation (equation 3.97) and Table 3.11, we can calculate the molecular weight (Mv) as follows:
[n] = K * Mv^a
Substituting the values:
2.73 x 10^-2 = K * (4.26 x 10^6)^a
Note that the units of K are irregular and usually omitted from tables, depending on the value of a.