Hello students, the question says that you have just assembled a volume ventilator and performed a leak test.
00:22
The cycling pressure achieved when the wave was occluded was only 55 centimeter water.
00:32
So the cycling pressure is was 55 centimeter water.
00:47
Now the question asked is describe the step -by -step procedure you would use to rectify the problem.
00:57
The first step is to identify a problem and the problem in this case is that the cycling pressure achieved when the wave was occluded was only 55 centimeter water h2o.
01:31
And the second step would be check for any obvious issue and inspect the volume ventilator for any visible signs of semi or malfunction and ensure that all connections are secure and that there are no leaks or blockage.
02:10
Third step would be to review the manufacturer's instruction and refer to the manufacturer's instructions or user manual for troubleshooting guidance specific to the volume ventilator model being used and follow any recommendation recommended steps or procedures provided.
02:55
And the fourth step would be to check setting and parameter and review the settings and parameters of the volume ventilator to ensure that they are correctly configured and we have to pay attention to the in -expiratory pressure, expiratory pressure, tidal volume setting.
03:33
Another step would be to perform a thorough leak test.
03:44
We have to repeat the leak test in order to ensure that there is or there are no leaks in the system and this can be done by occluding the wave and measuring the cycle pressure achieved.
03:56
If the cycling pressure is still below the desired level then we have to proceed to the next step and another step would be inspect and replace component.
04:19
If necessary we have to inspect the components of the volume ventilator such as tubing, connectors and valve for any signs of damage or wear and replace them if they are faulty or worn out.
04:45
Another step would be to we have to calibrate the volume ventilator.
04:53
If the cycling pressure is still not achieved within the desired range then we have to calibrate the volume ventilator according to the manufacturer's instructions and this may involve adjusting the pressure sensor or other setting...